Hello all!,
Can the CSI 2130 analyzer be configured to measure the gap voltage in Bentley Nevada systems?.
Thanks in advance.
 
Posts: 42 | Location: trinidad & tobago | Registered: 02 September 2007Reply With QuoteEdit or Delete MessageReport This Post
Posted Hide Post
For CSI 2120 Analyzer:
Select Sensor Type = DC
Set Sensitivity = 1.0

to get VDC measurement

Walt
 
Posts: 971 | Location: Massachusetts | Registered: 27 April 2005Reply With QuoteEdit or Delete MessageReport This Post
Posted Hide Post
Thanks
The the settings should be the same with the 2130 analyzer.
Where can I find the procedure for setting/ configuring the 2130 analyzer to measure the Gap voltage?
 
Posts: 42 | Location: trinidad & tobago | Registered: 02 September 2007Reply With QuoteEdit or Delete MessageReport This Post
Posted Hide Post
Vibhigh,

The only thing that I will add is that you need a Votage input adapter and you need to set Sensor Power = off. You can set up one or more Off-Route points or a Route within RBMware/AMS. If you have a Analysis Parameter Set for proximity probes, then you can have one Parameter set for "DC-Meas" so you get probe off-set in default units.

Walt
 
Posts: 971 | Location: Massachusetts | Registered: 27 April 2005Reply With QuoteEdit or Delete MessageReport This Post